A Ventilator Providing Pressure Support Ventilation to a Patient Is

Question 37

Multiple Choice

A ventilator providing pressure support ventilation to a patient is set to flow trigger on spontaneous breaths (base flow = 10 L/min; trigger = 3 L/min below base flow) . While at the bedside, you observe that the ventilator trigger rate is much higher than the patient's spontaneous rate, causing noticeable patient discomfort. Which of the following could be causing this problem?
I.circuit leaks
II.inadequate base flow
III.circuit movement


A) I and II
B) I and III
C) II and III
D) I, II, and III